how can i have one peak in melting curve of primer in Real-time PCR?
Relevant answer
Answer
Ritu Kumari is correct, the first thing to check is that you are amplifying only one PCR product.
But if all of your controls are working, and you know it's just one molecule, and you consistently see a double peak for ALL of your samples with that particular pair of primers, see my explanation below.
How large is your amplicon? The longer the DNA molecule, the more likely that it will melt in "regions" based on GC content. I've seen it when using melt-curve based genotyping. Anything over ~120 bp or DNA with regions of very different %GC will often get a messy peak.
  • asked a question related to Real-Time PCR
Question
3 answers
We aimed to compare mRNA expression of 2 genes in the human buccal epithelium quantified using the real-time PCR method in 2 time-point and in 2 groups. How I can calculate the minimum sample size? Whether is it enough per 10 patients per group (20 in general)? If yes, how ot can be proved.
Relevant answer
Answer
The following R-package could provide useful if you decide on using any of the following tests: two-proportions z-test, a two-group comparison of Poisson data,comparison of negative binomial data, one-sample or two-sample t-test, t-test for non-zerocorrelation, Fisher’s exact test, signed rank test, sign test, rank sum test, ANOVA, andsingle-predictor Cox regression (PDF) Computing Power and Sample Size for the False Discovery Rate in Multiple Applications. Available from: https://www.researchgate.net/publication/378807566_Computing_Power_and_Sample_Size_for_the_False_Discovery_Rate_in_Multiple_Applications [accessed Apr 25 2024].
  • asked a question related to Real-Time PCR
Question
7 answers
Hi ResearchGate community,
I have been trying to learn more about the optical differences between block-based real-time PCR machines like ABI StepOne versus rotor-based machines such as MIC or RotorGene systems.
I understand that some systems rely on ROX as a passive reference dye while others state that it is optional to incorporate it and others do not need such a factor at all.
My question is if you add this fluorescent dye to your master mix, would it interfere with the detection when it is being amplified using one of the systems that do not need such normalization?
Highly appreciate any insight in this regard.
Best,
Negar
Relevant answer
Answer
Dear ResearchGate community,
I'm somehow referring to the same question, I was wondering if it is fine to use a Sybr green master mix containing ROX for a machine that does not require ROX addition, The CFX96 C1000 touch from Biorad will this addition affect the signal detection and if yes are they any ways to subtract it? looking forward to your insights.
  • asked a question related to Real-Time PCR
Question
2 answers
Hello everyone!
I am using TaqMan probe-based Real-time PCR for the detection of typhoidal salmonella. Since I am not concerned about the expression level of gene, I just want to detect if there is any typhoidal DNA present in my sample or not. Do I still need to add internal control in my reactions?
If yes, then should I add the primer and probes for the housekeeping gene in every reaction since I am doing multiplex PCR or can I add it in a few reaction tubes in every run separately?
Relevant answer
Answer
Yes, it is recommended to include an internal control in your TaqMan probe-based Real-time PCR assay to monitor the performance of the assay, detect potential issues, and ensure the reliability of the results. You can either design the internal control to work in a multiplex PCR with your target DNA or run separate reactions for the target DNA and the internal control.
good luck,
Djihad Chenna
  • asked a question related to Real-Time PCR
Question
6 answers
Hello everybody.....
the concentration of my plasmid DNA is 126ng. i want to do real-time PCR taqman probe detection......i want to do serial dilution that should start from 10e8 copy number....so can anybody tell me how to dilute it (how much plasmid DNA and ddwater should be added to make it 10e8). i have attached the calculated values in the attachments. thank you
Relevant answer
Answer
Hello,
Titration of mixed solutions containing nitric acid (HNO3) and sulfuric acid (H2SO4) can be challenging due to the complex chemistry involved. However, there are methods available to accurately determine the concentrations of these acids in a mixture. One commonly used technique is the "back titration" method, which involves a series of chemical reactions to determine the acid concentrations step by step.
Here is a detailed and orderly explanation of the back titration method for titrating HNO3-H2SO4 mixtures:
  1. Dilution: To begin, you should dilute the mixture of HNO3 and H2SO4 with a known amount of deionized water. This dilution step is crucial to ensure that the concentrations are within the range that can be accurately measured using standard titration techniques.
  2. Titration of Excess Base: Add a solution of a strong base, typically sodium hydroxide (NaOH), of known concentration to the diluted acid mixture. The base will react with the excess sulfuric acid to form sodium sulfate (Na2SO4) and water (H2O). The reaction can be represented as follows:H2SO4 + 2NaOH → Na2SO4 + 2H2OThe reaction with nitric acid (HNO3) does not occur at this stage, as it is a weaker acid.
  3. Determination of Excess Base: Continue adding the NaOH solution until the excess base is completely neutralized, resulting in a pH jump. You can monitor the pH using a pH meter or pH indicator. At this point, all the H2SO4 has been converted to Na2SO4.
  4. Titration of Residual Nitric Acid: The remaining nitric acid (HNO3) in the solution is now titrated with a standard sodium hydroxide (NaOH) solution. The reaction between HNO3 and NaOH can be represented as:HNO3 + NaOH → NaNO3 + H2OThe endpoint of this titration is determined using a pH indicator that changes color at the pH at which the reaction is complete. Phenolphthalein is commonly used as an indicator in this case.
  5. Calculation: Calculate the concentrations of H2SO4 and HNO3 based on the volumes and concentrations of the NaOH solutions used in both titration steps. It is important to consider the stoichiometry of the reactions to determine the molarities of the acids accurately.
This back titration method allows you to determine the concentrations of both HNO3 and H2SO4 in the mixture. It is important to perform accurate measurements and titrations to obtain reliable results. Additionally, using standardized solutions and appropriate laboratory techniques is essential for the success of this titration method.
I hope this explanation helps you understand the procedure for titrating HNO3-H2SO4 mixtures. If you have any further questions or need clarification on any aspect of this method, please feel free to ask.
With this protocol list, we might find more ways to solve this problem.
  • asked a question related to Real-Time PCR
Question
5 answers
I recently conducted cDNA synthesis followed by conventional PCR for quality assessment. In the course of this experiment, I observed some unexpected results that have raised questions about the reliability of my cDNA and its potential impact on real-time PCR experiments.
Specifically, here are the key observations:
  1. RT- Sample: The reverse transcription negative control (RT-) showed the presence of two bands, with one of them being my target gene. This was unexpected as the RT- control is typically used to confirm the absence of cDNA synthesis. I'm puzzled by the presence of my target in this control (the NTC did not show any bands).
  2. Sample Variability: In the PCR results, all of my samples indicated my target gene, but I noticed variations in both stringency and intensity of the bands, despite matching the RNA amounts to 1000 ng during cDNA synthesis (The A260/A280 of all the samples were in the range of 1.8-2). This variability across samples is concerning and may affect the reliability of my results (The A260/A280 of all the samples were in the range of 1.8-2).
My primary concern is whether these observations in the cDNA synthesis and conventional PCR could potentially impact the outcomes of my real-time PCR experiments. Real-time PCR requires high precision, and any issues with cDNA quality or PCR variability could affect the accuracy of gene expression quantification.
I would greatly appreciate insights from the research community regarding the possible reasons for these observations and their potential implications for downstream real-time PCR experiments. Your expertise and suggestions on troubleshooting or optimizing this process would be invaluable.
Relevant answer
not exactly.
  • asked a question related to Real-Time PCR
Question
3 answers
I want to synthesize Gene-specific real-time PCR primers for my differentially expressed novel and conserved micro RNAs in plants.
Relevant answer
Answer
You are welcome
  • asked a question related to Real-Time PCR
Question
3 answers
helo
Relevant answer
Answer
Lets look at definitions and applications of the methods in your question:
The western blot (also known as western blotting), is a widely used analytical technique in molecular biology and immunogenetics to detect specific proteins in a sample of tissue homogenate or extract (or to detect a specific protein in a blood sample for instance).
It is noteworthy to mention that ELISA is a more simple and faster procedure than Western blotting, which is less specific. Western Blotting is a highly successful testing method for confirming positive results from ELISA tests. It is also used as a confirmatory test as it is difficult to perform and requires a high skill level.
On the other hand, real-time polymerase chain reaction (PCR) is a technique of molecular biology based on the polymerase chain reaction. It monitors the amplification of a targeted DNA molecule during the procedure (in real time), not at its end, as is the case in the conventional PCR. Real-time polymerase chain reaction can be used quantitatively and semi-quantitatively.
  • asked a question related to Real-Time PCR
Question
12 answers
Hi, I'm doing presence/absence experiments with a comercial kit qPCR. The software requires an IPC-blocking agent but the kit does not have. Is there any configuration that allows it without IPC blocking agent? Moreover, the presence/absence graph does not appear in the analysis.Who know how to fix this problem? Thanks.
Relevant answer
Answer
Hi Erika Perez, I’m experiencing the same issue. Have you found a way to solve it?
  • asked a question related to Real-Time PCR
Question
3 answers
I extracted RNA from the animal tissues by Trizol and phenol technique. The A260/A230 and 260/280 ratio were sometime lower or higher than a normal range. Do this make problems for my real-time PCR results?
Relevant answer
Answer
Hi Parisa,
As per my understanding low or high RNA purity and concentration ratios can interfere with reverse transcription and PCR amplification, leading to inaccurate results. Clean up your RNA before performing real-time PCR.
Best of luck!!
  • asked a question related to Real-Time PCR
Question
5 answers
I am trying to find RCN in real-time PCR on SYBR Green, but Cq among 3 repetition of the same DNA is going up and down around 1.2 cycle, so the error is too large. I'm working on genomic DNA extracted from buccal swabs, and have no way to measure its concentration, but I centrifuged the DNA before testing, diluted the DNA 100 times to reduce the concentration, but Cq continues to jump. Is there any way to handle this?
Relevant answer
Answer
Thanks for the update. Under optimum conditions ("perfect" amplification efficiency), the Ct should shift by about 3.3 cycles per 10-fold dilution (=log2(10)), hence about 6.6 cycles per 100-fold dilution. The shift will be larger under suboptimal conditions). If your normal samples have Ct-calues of 24 or larger, then it should not happen that 100-fold dilutions of these samples have Ct-values of less than 30 (you mentioned 27). This is strange.
Apart from this I don't see anything obvious that would possibly explain your findings. Does this happen only for a specific primer/assay or do you see this also for other primers? If yes, it might be due to the instrument. If only a particular set of primers/assay doesn't work well you might just design and order (sligtly) different primers, what might solve your problem.
  • asked a question related to Real-Time PCR
Question
2 answers
Hello colleagues, I found the Eco system real time PCR illumina device in the lab. The last people who used the device are already out... So after the initial excitation about using it, I discovered that the sample sizes are quite smaller than the normal PCR samples I used to make in Eppendorf devices. So is the real-time PCR sample preparation the same for this device as any other PCR? or is it different? any and all information is highly welcome, I went over the user guide several times, it generously explains how to use the device, except for the part pertaining to the actual sample preparation. Thanks a million in advance
Relevant answer
Answer
The sample preparation for real-time PCR on the Illumina Eco system does follow the same general principles as standard real-time PCR. However, there are some specific requirements to adapt the protocol for the Eco system:
- Reaction volume is very small - only 10-20 μL per well compared to 20-50 μL for standard PCR plates. So reagents need to be scaled down accordingly.
- Special 384-well plates are used. Make sure to use compatible optical plates provided by Illumina.
- More concentrated primer stocks may be needed to accommodate the smaller reaction volumes. 100 μM primer stocks are common.
- Always include positive and negative controls.
The actual real-time PCR reagents (Taq polymerase, buffers, dNTPs) and cycling conditions can remain the same as standard PCR. Just the preparation needs to be adapted for the smaller 384-well Eco plates.
  • asked a question related to Real-Time PCR
Question
2 answers
Can anyone comment or suggest how to standardize bacterial 16S realtime PCR primers for detecting bacterial presence/absence through qPCR?
Relevant answer
Answer
Hi Robert, Thanks for the kind reply.
  • asked a question related to Real-Time PCR
Question
3 answers
I am conducting a comparative experiment using real-time PCR to determine the threshold cycle (Ct) values. For one of the gene primers, I have established that the ideal annealing time is 10 seconds. When I increase the annealing time beyond 10 seconds, I observe no amplification. I have tried various standardization methods and have decided to continue with the 10-second annealing time. However, with this setup, I am obtaining Ct values but no visible amplification, possibly due to insufficient elongation time. I am using Applied Biosystems SYBR Green Master Mix.
Given this situation, I am uncertain whether I should conclude the experiment based on the obtained Ct values, considering the absence of visible amplification. Is it okay to proceed with the obtained Ct values? Your insights and guidance regarding this matter would be greatly appreciated. Thank you.
Relevant answer
Answer
No, you can not use Ct values that don't indicate that your gene-of-interest was amplified.
Something is not working correctly. Either you have primer dimers, your qPCR product is too large, you have contamination, the threshold is set too low in your analysis, or some combination of the above.
What do you see on your standard curve and your positive controls?
  • asked a question related to Real-Time PCR
Question
4 answers
Hello, I am bachelor student. I extract DNA from urine sample using two different DNA extraction kits and detect alpha thalassemia 1, SEA deletion type by real-time PCR. I use the same of pcr mixture and pcr condition but Melt peak aren’t same location. in the picture, blue is dna from kit A, green from kit B, pink is control and red is NTC. the dna from kit A have melting temperature lower than DNA from kit B. So, l’m not sure what happen.
DNA purity affects to shifted or not?
Does anyone have recommendation for this issue?
thank you
Relevant answer
Answer
Many thanks for taking the time to give me more details.
Francesc Codony's suggestion might make the situation clearer, I guess.
Best,
N
  • asked a question related to Real-Time PCR
Question
4 answers
Hi, has anyone tried to sequence real-time PCR amplicons using the Flongle?
Relevant answer
Answer
Yep, it can be done. Though most of our experience has been on the Minion, I know we've done some work on the Flongle too.
  • asked a question related to Real-Time PCR
Question
9 answers
Hi there!
I got amplification of my genes in the negative control in Real-time PCR experiment although the Ct values were above 30. I setup my reaction again after making new dilutions of PCR reagents, cleaning the workbenches, using unopened autoclaved tips, and cleaning my pipettes with 70% ethanol. I tried to remove every possible source of contamination but still I am getting the same Ct values in my negative control. It is a TaqMan probe-based multiplex PCR reaction where I am amplifying three genes in the same reaction and I am getting amplification of two of these genes in my negative control. Can anyone guide me how to get rid of this amplification?
Relevant answer
Answer
I recommend verifying these steps:
· Cross-contamination: Take care to prevent cross-contamination between samples by using separate equipment and workspace for the negative control. Double-check that pipettes and consumables are not shared.
· Contaminated reagents: To minimize the risk of contamination, handle reagents with caution. Use fresh reagent stocks from a different batch, store them properly, and ensure they are not compromised.
· Aerosol contamination: Employ filtered pipette tips, work in a clean environment, and keep PCR tubes or plates covered while performing actions like opening tubes, vertexing, or pipetting to minimize the potential for airborne DNA fragment contamination.
· Carryover contamination: Thoroughly clean and decontaminate laboratory equipment, including pipettes, tubes, and the thermal cycler, to eliminate any residual DNA from previous experiments that could contribute to carryover contamination.
My suggestions:
1. Repeat the experiment using new reagents, preferably from a different batch, to eliminate the possibility of contaminated reagents.
2. Set up a fresh negative control in a separate area using different pipettes and consumables to reduce the risk of cross-contamination.
3. Conduct additional negative controls using sterile water instead of template DNA to determine if the issue persists. This will help identify if the contamination originates from the reagents or the laboratory environment.
4. Ensure that PCR tubes or plates are properly sealed to prevent aerosol contamination during the experiment.
  • asked a question related to Real-Time PCR
Question
1 answer
How to evaluate the quantity of gut microbiota through real-time PCR?
Relevant answer
Answer
I heard Qiagen has a kit. I was reading about it last week. Please check the files attached.
Thanks
  • asked a question related to Real-Time PCR
Question
3 answers
when Real-time PCR consider a high throughput technique? Is gene expression analysis by qPCR give a low throughput data?
Relevant answer
Answer
thank you
  • asked a question related to Real-Time PCR
Question
1 answer
Hi everyone. I'm making a real-time PCR primer efficiency curve of long non-coding RNA of exosome, but the curves are inverted. In concentrated samples, Cq is higher (above 30) and in more diluted samples we have lower Cq (above 26 and 27). I would like to know if someone has already made an efficiency curve with molecules from exosomes and how it was because I can't find a solution.
Relevant answer
Answer
Hi Suellen,
I'd suggest you look at your trace data. I believe you'll find the curves are not at the quality as required to draw quantitative results from a standard curve. I'd suggest you look into the following points which may or may not apply to your work (as there are many variabilities/requirements of qPCR):
1. Use a stable DNA source of control within a vector
2. Dilute your controls and samples to be a maximum of 10% of your qPCR reaction
I've worked with EV RNA/DNA many times and have not had any issues differing from the norm when it comes to qPCR. I'd suggest you perhaps chat with someone from your group/laboratory and go over the quality of your sample/methodology to confirm.
Good luck.
  • asked a question related to Real-Time PCR
Question
2 answers
Which tools or methods are used to compare the relative gene expression of real-time PCR value other than the Livak method??
I have values of CQ of both gene and internal reference. I want to know which methods compare the gene expression among the samples. The relative expression between samples.
I have used the Livak method. But I want to know other methods or tools.
Please guide me.
Thank you
Relevant answer
Answer
Hello! I like to use the "qpcR" R software to fit curves, but it isn't necessary. you can use the "delta delta Ct" method. Here is the gold standard instructions for qPCR experiments. Good Luck! https://www.gene-quantification.de/national-measurement-system-qpcr-guide.pdf
  • asked a question related to Real-Time PCR
Question
6 answers
Hi, I have NTC contamination in real-time PCR. I checked all suspicious items such as primers, master mix, and distilled water, and also the workspace. but unfortunately, I can not remove the contamination.
Relevant answer
Answer
Hi Zeynab,
For such issues, you can simply contact Dr. Shanehbandi at Tabriz Medical University, an expert in PCR and troubleshooting. 09144160144
All the best,
Daniel
  • asked a question related to Real-Time PCR
Question
3 answers
What is wrong with my real-time PCR results?
Relevant answer
Answer
Check out the scale on the Y-axis on both graphs. You have minimal amplification from any of the samples, that's why the data look so messy.
Start with just a positive & negative control. Once you get those working, then add in the standard curve samples.
Good luck!
  • asked a question related to Real-Time PCR
Question
1 answer
I need to do some analysis on my personal computer and unlike the Quantstudio software, I can't find a free version of the 7500 Real-Time PCR software
  • asked a question related to Real-Time PCR
Question
10 answers
Hi,
I'm analyzing the same DNA samples using conventional PCR and real-time PCR (Sybrgreen), with the same PCR parameters and primers. However, while in the qPCR most of the samples are negative, in the conventional PCR they are mostly positive. Does anyone has an idea how to explain that?
Thanks!
Relevant answer
Answer
There are several reasons why conventional PCR and real-time PCR (qPCR) can yield different results, even when using the same PCR parameters and primers. Here are some possible explanations:
  1. Sensitivity: qPCR is generally considered to be more sensitive than conventional PCR because it can detect and quantify PCR products in real-time as they accumulate. Conventional PCR relies on endpoint detection, which may be less sensitive, especially if the PCR product is present at low levels.
  2. Contamination: Real-time PCR is more susceptible to contamination because it involves handling of PCR products in real-time. Therefore, it is important to use appropriate precautions and controls to prevent contamination. Contamination can lead to false-positive results in conventional PCR, while qPCR may be able to detect and exclude such results.
  3. PCR inhibitors: Real-time PCR is more sensitive to PCR inhibitors that may be present in the sample, such as hemoglobin or other substances that can co-purify with DNA. If present, such inhibitors can affect the efficiency of PCR amplification, leading to false-negative results in qPCR.
  4. Differences in primer specificity: Although the same primers are used in both conventional and real-time PCR, the specificity of the primers may differ between the two methods. This can lead to differences in PCR product yield, with one method producing more non-specific products than the other.
  5. Differences in reaction conditions: Although the same PCR parameters are used, slight differences in reaction conditions, such as the annealing temperature, extension time, or MgCl2 concentration, may affect the efficiency of PCR amplification and lead to different results.
Several factors can contribute to different results in conventional PCR and real-time PCR, including differences in sensitivity, contamination, PCR inhibitors, primer specificity, and reaction conditions. It may be helpful to optimize the PCR conditions and controls to minimize these sources of variation and improve the accuracy and reproducibility of the results.
These video playlists might be helpful to you:
  • asked a question related to Real-Time PCR
Question
3 answers
We performed qPCR analysis using SybrGreen dye, where the template was gDNA, on the Applied Biosystems 7500 Fast Real-Time PCR System (software ver 2.3). The reaction conditions were optimized using standard PCR conditions (denaturation at 95°C for 60s, annealing at 58°C for 30s, elongation at 72°C for 30s). First, these conditions were tested. We also tested the default program for SybrGreen (2 step, 60°C annealing + melting curve, photo1,2) and many other combinations each time, without obtaining any curves (photo3). I know that the reaction itself, reagents and temperature profile is well optimized because after the reaction, we made an electrophoresis of the product (photo4).
Do anyone have experience with this software? Is this software bug (do you press any extra option?) or do I make something wrong? Additionally, is it possible to preview the amplification plot such that it is presented on a linear and not a logarithmic scale during reaction? I will be grateful for any tips.
Relevant answer
Answer
This could happen if the machine is reading the ROX value from your ctr.
You need to go to "set up" then "assign targets to wells", for the endogenous ctr. select "none" in the reference. Then click on "analyze". Your amplification curves should appear.
If again no amplification, then you need to increase the amount of your starting cDNA samples or increase the number of cycles to 45
+
choose the delta-delta-CT option from your second picture as mentioned by Yi-Chun Cheng
  • asked a question related to Real-Time PCR
Question
4 answers
My concentration was 123.5 ng/uL
260/280 Ratio - 1.725
260/230 Ratio - 0.401
Can I synthesize cDNA from the above RNA?
Need suggestions please.
Relevant answer
Answer
For cDNA synthesis, high-quality RNA with an A260/A280 ratio of 1.8 to 2.2 is ideal, and a minimum concentration of 1 μg/µL is usually recommended. For Real-Time PCR, the optimal RNA concentration is highly dependent on the gene of interest, and can range from 1 ng/µL to 500 ng/µL.
  • asked a question related to Real-Time PCR
Question
3 answers
I know that in LighterCycler 480 system, we can use Absolute quantification/fixed points methods and then set a reasonable noise band to estimate the CT values.
My workmate told me to use this analysis method in repeating her work, however, the qPCR machine in our lab is Applied Biosystems ViiA ™ 7 Real-Time PCR System. And I really can not find out how to set the noise band or apply this method.
Thank you so much for your attention and advices!
Relevant answer
Answer
Oh I finally figured out it, thank you so much.
  • asked a question related to Real-Time PCR
Question
2 answers
What factors affect melt peak [-d(RFU)/dT) in real-time PCR?
What should I do if I want to increase my melt peak [-d(RFU)/dT)?
Relevant answer
Answer
Real-Time PCR (Probe type) melt peak factor, also known as the Tm (melting temperature) or Tm peak, is a measure of the stability of a double-stranded DNA molecule. It is the temperature at which half of the double-stranded DNA dissociates into single strands. In real-time PCR, the Tm peak factor is used to determine the specificity of the PCR reaction.
The Tm peak factor is used to determine the specificity of the PCR reaction because it is affected by the length and nucleotide sequence of the DNA fragment, as well as the buffer conditions and the presence of other DNA sequences in the reaction mixture. A specific and efficient PCR reaction will produce a single, sharp Tm peak, while a non-specific reaction will produce multiple, broad Tm peaks. The Tm peak factor is therefore a useful tool for assessing the specificity and efficiency of real-time PCR reactions.
  • asked a question related to Real-Time PCR
Question
3 answers
Why products with more than 200 base pairs are not amplified in real-time PCR? Are there any images of this?
Relevant answer
Real-time polymerase chain reaction (real-time PCR) is a widely used method for amplifying and quantifying short DNA fragments, typically less than 200 base pairs (bp) in length. This method is based on the simultaneous amplification and detection of DNA during the PCR reaction, and it is commonly used in a variety of applications, including gene expression analysis, pathogen detection, and genetic typing.
Products larger than 200 bp are typically not amplified effectively in real-time PCR for several reasons:
  1. PCR specificity: Real-time PCR is based on the selective amplification of a specific target DNA region using specific primers. Longer DNA fragments are more likely to contain non-specific primer binding sites, leading to the amplification of non-target DNA.
  2. Efficiency of amplification: Real-time PCR is designed to amplify small DNA fragments very rapidly, and the rate of amplification decreases as the length of the DNA fragment increases. This can lead to decreased sensitivity and increased variability in the results when amplifying larger DNA fragments.
  3. Detection limitations: Real-time PCR uses fluorescent dyes or probes to detect the amplified DNA during the reaction. These dyes and probes are optimized for short DNA fragments, and their performance may be reduced when detecting longer fragments.
For these reasons, it is generally not recommended to use real-time PCR to amplify products larger than 200 bp. If longer fragments need to be amplified, alternative methods such as standard PCR or long-range PCR may be more appropriate.
  • asked a question related to Real-Time PCR
  • asked a question related to Real-Time PCR
Question
7 answers
I performed real-time PCR on cDNA of samples that had a nanodrop quantity of 1000+ ng/ul, but when i perform i do not get any proper results. i am attaching the report of my real-time. kindly help me out. should i synthesize the cDNA again or should i change something in the way i'm performing real-time pcr.
Relevant answer
Answer
If the power went out DURING your cDNA synthesis, then the most likely explanation is that you didn't synthesis any/enough cDNA.
The nano drop will only tell you the concentration of nucleic acids - not the type of nucleotide, length of molecules etc.
Try again and good luck!
  • asked a question related to Real-Time PCR
Question
4 answers
Hi there!
I am on the way to order a Real-Time PCR system for following analysis in my lab,
  • Relative quantification of gene expression
  • Genotyping by HRM
  • MicroRNA expression
I‘ve used Rotor Gene Q before, but I need plate-based system. So I would like to switch Applied Biosystems devices.
I am confused between Stepone plus and Quantstudio 3 instruments.
I need your recommendations with reasons.
P.S. I am a Mac user. Software compatibility is important.
Cheers
Relevant answer
Answer
Applied doesn't have a native mac version so for both you will have to use a virtual windows machine on your computer.
It have been a while but BioRad used to have a mac version for their qPCR software and I quite liked the CFX when I worked with it (nearly 10 years ago).
  • asked a question related to Real-Time PCR
Question
4 answers
Hello everyone,
I'm new to Real-Time PCR and I'd like your insight on this. I have generated standard curve by doing 1:10 serial dilution in triplicate. However, the result obtained, of which R value is to determine the efficiency of the assay is poor (R> 114%) and there are only three points out of 10 that fall on the line. Should I rerun test on the rest of the deviated points or repeat the whole procedure again?
Many thanks
Relevant answer
Answer
You'll have to repeat all of it.
But it's normal to have limits of your useful detection range. Too few copies and you don't get efficient amplification. Same can happen for too many copies of the template.
The R-square value tells you have precise & accurate repeatability for technical replicates.
The slope of the Ct vs template concentration tells you the reaction efficiency.
Both of these measures are important.
Good luck!
  • asked a question related to Real-Time PCR
Question
2 answers
Hi, I conducted qPCR using probe included FAM and BHQ-1.
At two annealing temperatures(60 and 65 C), the Ct value, R^2 and PCR efficiency in the amplification plot were also within the normal range.
When the PCR product was confirmed on an agarose gel, a normal band was confirmed.
However, only a slightly higher level of fluorescence than background was detected in the multicomponent plot(square box in figure).
But why is the Ct value calculated? Can I continue to use this probe?
In the same experiment, graphs for other probes containing FAM were normally generated.
Relevant answer
Answer
Is it probe degradation? If the probes degrade, and the quencher probe separates from the reporter, it will cause phosphorescence that increases linearly instead of exponentially. Remember that automatically these graphs are shown on a log scale. I do not believe there is a true asymptote in your left hand plots (directed to OP). You can linearize the graph (change axis so it is not a log scale) and see if the growth is exponential or linear. If it is not exponential, it is not caused by amplification but rather by probe degradation.
To OP: It's hard to read your axes, but it seams your c(t) is very different for your plot on left and plot on right. If you made them the same, and scaled them the same, the two graphs would not look the same (I think, if I'm deciphering the axes correctly, but hard to read). If you placed your C(t) at 0.1, the plot on left would look very unreliable, the plateau being far too low.
  • asked a question related to Real-Time PCR
Question
4 answers
I used a KOD DNA polymerase for PCR that works very well. Then I used this enzyme to run a TaqMan probe real-time PCR but the results waere unexpected. I attached the results. The fluorescence rise very fast during initial cycles and it reach to a plateau much faster than the control. The control enzyme is a Hot-start but the KOD is not. Can this be the cause of this phenomenon or is there another reason involved?
figure legend:
Red: Control enzyme (Hot-start)
Blue, green, purple, and pink: KOD with different buffer.
Relevant answer
Answer
Hi
Yes, I also agree you need to optimize the reactions first, what ever polymerase you are using. Also the nature of enzyme (hot start/regular) will also make difference.
  • asked a question related to Real-Time PCR
Question
4 answers
What would happen if my primer pair can amplify several genes? (ie. it is not specific to only one gene) Would it be the cause of false positive results?
Additional info.
1. I have BLAST my primer pair and it can bind only my bacterial species of interest.
2. I use a specific nucleotide probe for fluorescence detection in real-time PCR.
Relevant answer
Answer
On top of what Aissa Saidi said, I would say that it could also effect sensitivity as if the primer for target is being producing other amplicons. The test will not be able to have enough primers/dNTPs etc for your target of interest if you have templates competing for the primer. Hence, even if your the target template was there, and you had some primer-target interaction, the non-specific amplicons could out compete the target of interest amplicons so you would always have sensitivity issues in this case.
  • asked a question related to Real-Time PCR
Question
2 answers
Hello everyone,
which genes can I check in real-time PCR in cardiac fibrosis? Has anyone done it? I would appreciate it if you could help me.
Relevant answer
Answer
Pankaj Kumar Thanks a world
  • asked a question related to Real-Time PCR
Question
2 answers
I have made real time pcr with B-ACTIN and TNF-ALPHA . It give ct values with B actin but not give any values with TNF-alpha.
I use 25microlitre pcr
1.50 microlitre primer
3 microlitre cdna
7.50 microlitre cybergreen
13 microlitre water free
Relevant answer
Answer
Hi,
3 microlitre cdna, this is too much.. I would suggest to check the Ct values for B-actin first. Check your primers too.
  • asked a question related to Real-Time PCR
Question
15 answers
I'm passionate in oncology research. I did bachelor in Zoology and now pursuing MPhil Molecular Biology and Biotechnology and working on cancer genetics. I am greatly interested to do PhD in cancer research from a world renowned institute but I think with this profile I would get a position in top ranked institute for PhD. Should I go for another master from a renowned foreign institute with major in oncology?
Thanks
Relevant answer
Answer
I am also at a similar position as you are in. I have completed my MSc. in Genomic Medicine from the University of Birmingham. I am planning to learn Bioinformatics, maybe I will do that from 6months courses and then try to gain some experience in this field.
I think gaining some practical experience is quite necessary at the moment for you as well. After that you can always opt for PhD.
  • asked a question related to Real-Time PCR
Question
2 answers
Hi everyone
Please give me a solution about this question.
Thanks a lot
Relevant answer
Answer
Thanks a lot for your answer
  • asked a question related to Real-Time PCR
Question
4 answers
Hello, our Real-Time PCR system (AB7500) is acting up with the status constantly showing “Waiting for the heated cover to reach temperature” when we start the run. Neither the cover, nor the block are getting heated. Can anyone suggest what the error might be and how this might be fixed?
Thanks in advance!
Relevant answer
Answer
Better check your service contract to see if your machine is under warranty. You will not be able to fix this yourself.
Good luck.
  • asked a question related to Real-Time PCR
Question
1 answer
Of course, the samples' Cts do not fall within the standard curve.
Just curious about the scientific community's impressions on this topic because we recently discussed it in our lab.
Relevant answer
Answer
ct values above 40 are not considered as the fine category values to calculate the relative mRNA expressions. usually, when we set the qPCR, the maximum cycle we set is 40. In order to get the desired cycle ratio; one way to do this is by increasing the DNA concentration.
  • asked a question related to Real-Time PCR
Question
3 answers
The expression of nirS and narH genes involved in the nitrite reductase and nitrate reductase pathways of Thiobacillus denitrificans ATCC 23644 treated with zero-valent iron nanoparticles was measured by Real-time PCR based on SYBR-Green I fluorescence, which showed a significant increase compared to the control group.
Relevant answer
Answer
This step is vital to double check of the absence of primer dimer and the specificity of the used primers. In addition to avoid the background noise of the dye used. Therefore, it has become an important step particularly when using SYBR Green I. Thus, it has been recommended to conduct a melting curve analysis.
  • asked a question related to Real-Time PCR
Question
2 answers
I performed a Real-time PCR on colon cancer cell lines after treatment with some anti-cancer agents. Although the expressions of caspase 8, 9 and P53 are increased but the Bcl2 expression is increased as well. How is it possible? because Bcl2 is an anti-apoptotic protein.
Relevant answer
Answer
No problem. Happy to be.
Best reagrds
AB Bayazid
  • asked a question related to Real-Time PCR
Question
6 answers
Dear friends!
I am almost a newbie in the miRNA world :)
Try to identify miRNA in heart tissue samples according to a TaqMan™ Small RNA Assays user guide (No 4364031). I use Taqman chemistry, 7500 Fast Real-time PCR (Applied Biosystems).
And it didn`t work (please, see attached pics).
Do you have ideas why?
First pic. - RNA was not standardized by concentration, second - 5 ng of total RNA was added to RT mix. And things only went worse ((
I use TaqMan MicroRNA RT Kit, Universal PCR Master Mix, MicroRNA Assays (U6B, miRNA-1, -34a, 320 - all "hsa").
I would be very grateful for your help!
Have a good day!
Glory to Ukraine! Glory to the Heroes!
Relevant answer
Answer
Saurabh Mandal you were absolutely right. For unexplained reason,the TaqMan master mix didn`t work! We used another and got results.
Thanks!
  • asked a question related to Real-Time PCR
Question
2 answers
i have been developed a taq man real time pcr for my work… but i have a problem in maintaining my STANDARDs and after a week the load of plasmids gets low… how can i solve this problem?
Relevant answer
Answer
It may be that your dna is absorbing onto the plastic of the tube or it is being degraded by nucleases or acid pH.
You can store lyophilised and make up in TE just before the assay or add some non amplifying dna or rna to your standards before storage. The cold dna will block the absorption sites on the plastic tubes. Siliconisation of the storage tubes may also help
  • asked a question related to Real-Time PCR
Question
3 answers
Anyone who can help me in obtaining a protocol to detect mutation in pfdhfr and pfdhps genes using Real-time PCR.
Relevant answer
Answer
Hello dear Ali
PCR for detection
and
q-PCR for gene expression rate.
  • asked a question related to Real-Time PCR
Question
3 answers
Dear researchers
I do use multiplex qPCR, I use 4 different templates, Among these samples, only one sample give me one good Peak in flouresent per cycle and ct, everything is the same but the type of template. Does anyone could help me with this kind of abnormal peaks.‌
Relevant answer
Answer
As far as I know abnormal peak like represent Noise. This may be due to any contamination or your template is not pure due to which you are not getting desired peaks. usually, you see this kind of peaks before the reaction reaches ct value.
As you are using multiple primers have you checked for the formation of dimers?
  • asked a question related to Real-Time PCR
Question
5 answers
I am looking for to buy a RT-PCR machine-96 wells. Please mention your experience with the machine in your lab. Any pros and cons. Thanks
Relevant answer
Answer
Lightcycler (Roche) outperformes CFX.
  • asked a question related to Real-Time PCR
Question
3 answers
Dear ResearchGate community,
I am working on DNA samples obtained from patients screened for cervical cancer/ HPV infection. To calculate the viral load of each sample, I am looking for a technique to know the amount of the viral DNA present in each of the samples.
That said, considering the fact that the viral genetic content can be present both as episomes as well as integrated into the human genome, what would be the best approach to get insight from the "effective" viral load?
In other words, for the diagnostic purposes which one of the two types of viral DNA- if not both- is of significance, and how can it be quantified?
Many thanks in advance,
Relevant answer
Answer
Why don't you run Real-Time PCR (LightCycler)? It can be used for quantitative analysis of a gene copy number so it can be used for viruses as well.
  • asked a question related to Real-Time PCR
Question
6 answers
I have loaded my qPCR products onto a 2% agarose gel and ran them at 60V for 1 hour. I ran 2.5ng of cDNA in a 10uL qPCR reaction. As you can see in my image there is smearing of all bands in the heavy direction. I have tried reducing the amount of sample loaded and switching to TBE buffer and this helped slightly but still gives the smearing that is shown in the attached image. What else can cause this?
Relevant answer
Answer
You can quantify your obtained DNA by checking the concentrations. Concentrations lower than 1.7 might not give the desired bands.
  • asked a question related to Real-Time PCR
Question
3 answers
So I got these weird amplification curves in SYBR green qPCR using cDNA extracted mouse tissues. The miRNA was retrotranscripted with stem-loop primers. The curves are weird because the fluorescence was not uniform at the start of the reaction, increased suddenly and dropped after reaching plateau. The weirdest part is that the anomalies did not happen in every well, but only some of them. The melting curves showed one main peak and much fluctuation. However, another reaction done using the same RNA material and kit showed no anomaly. Sincere thanks in advance for your advice.
Relevant answer
Answer
You should contact tech support for your machine. Looks like an error in the hardware or software.
  • asked a question related to Real-Time PCR
Question
3 answers
I understand that the fragment size is a limiting factor, but I need to test and know what I can adjust in the cycle.
Relevant answer
Answer
Also, make sure that your primers are annealing to exons, not introns, or you won't get amplification from cDNA.
You can change the time & temperature for all 3 steps (denature, anneal, extend).
As Laura Morais Nascimento Silva says, check the ideal cycling conditions for your qPCR mix. Good place to start.
Not all primer pairs are efficient enough for qPCR, but in my experience, the primers that give a bright band at a range of annealing temps are ones that are more likely to work.
Good luck!
  • asked a question related to Real-Time PCR
Question
18 answers
Hello,
I have 2 groups in my study (treatment & control). I did taqman gene expression assays for the gene of interest (GOI) and a house keeping gene for each individual in the study Post treatment. I don’t have baseline values for any of the individuals. Now I have Ct values for both GOI & housekeeping gene for each individual in treatment & control groups. Can anyone help me in the next step?
I think the delta delta Ct method (Livak method) cannot be applied for my data as they are 2 different groups, not pre and post-treatment data. What can I use alternatively?
Thanks in advance.
Relevant answer
Answer
Manar Serageldin The two most common methods for analyzing qPCR data are double delta Ct analysis and the relative standard curve approach (Pfaffl method). Because both approaches make assumptions and have limits, the method you should employ for your analysis will be determined by the experimental design.
Take a look:
  • asked a question related to Real-Time PCR
Question
9 answers
While Working in molecular biology Lab, QuantStudio™ 3 Real-Time PCR System, 96-well, 0.2 mL, laptop is our tool.
I want to calculate the Genomic equivalence, then want to calculate it for parasite number per micro liter of blood.
Hence, I am looking advises from any one working on this area is very helpful.
Abdissa B.
Relevant answer
Answer
The best way to do this is to prepare a sample of parasite DNA from a known number of parasites, make a dilution series and then use this as a standard curve. If you cannot obtain such a sample, there are ways round the problem. I work with pathogens that cannot be cultivated in cell-free media and which are difficult or impossible to enumerate. To prepare my standards, I use a plasmid containing the target sequence (several suppliers can produce such plasmids using a synthetic sequence; I use Genscript). DNA concentration can be converted to number of copies by dividing by the mass of the plasmid. Then you make a dilution for use in the standard curve.
The problem with this approach is that you have to know (a) the number of copies of the target sequence in the parasite genome and (b) the number of cells in a parasite. If you don't know this, and you cannot get an accurately counted sample of parasites, then absolute quantitation is essentially impossible.
  • asked a question related to Real-Time PCR
Question
8 answers
What are the limitations and disadvantages of Real-Time PCR (RT-PCR)?
What is a more specific and sensitive technique that can be used in the laboratory instead, particularly in cancer diagnosis?
Relevant answer
Answer
The limitations of Real-Time PCR are as follows:
1. High false-negative due to presence of amplification inhibitors. This holds true for pathogen detection particularly, for new emerging or highly variable pathogen.
2. The multiplexing is still limited in Real-time PCR.
3. The variation increases with the number of cycles.
4. The overlap of emission spectra.
5. The occurrence of non-specific binding particularly, when carrying out SYBR green analysis.
6. The PCR product increases exponentially but one cannot monitor the amplicon size.
7. Kits are not available for all kinds of genes and disorders.
8. It needs specialized laboratories, operated by highly trained technicians for developing novel qPCR assays, which makes it too costly and difficult to scale up.
You could use other techniques that could be performed in the laboratory.
1) Fluorescence In Situ Hybridization (FISH), also known as molecular cytogenetic testing, is a way to visualize and document the location of genetic material, including specific genes or DNA sequences within genes. FISH is used to look for the presence, absence, relative positioning, and/or number of specific DNA segments under a fluorescence microscope. FISH is particularly helpful in identifying copy number variations, especially translocation and amplifications.
2) Comparative genomic hybridization (CGH) is a special FISH technique (dual probes) which is applied for detecting all genomic imbalances. CGH is used to determine copy number alterations of genome in cancer and those cells whose karyotype is hard or impossible to prepare or analyze.
3) Single Strand Conformational Polymorphism (SSCP) is one of the simplest screening techniques used for detecting unknown mutations (microlesions) such as unknown single-base substitutions, small deletions, small insertions, or micro inversions. A DNA variation causes alterations in the conformation of denatured DNA fragments during migration within gel electrophoresis. The logic is comparison of the altered migration of denatured wild-type and mutant fragments during gel electrophoresis.
4) Heteroduplex analysis in which a mixture of wild-type and mutant DNA molecules is denatured and renatured to produce heteroduplices. Homoduplices and heteroduplices show different electrophoretic mobilities through nondenaturing polyacrylamide gels.
5) Denaturing Gradient Gel Electrophoresis (DGGE) has been used for screening of unknown point mutations. It is based on differences in the melting behavior of small DNA fragments (200-700 bp); even a single base substitution can cause such a difference.
6) DNA microarrays can detect thousands of genes at once. DNA “chips” or microarrays can be used to test for multiple mutations. In this technology, single DNA strands including sequences of different targets are fixed to a solid support in an array format. On the other hand, the sample DNA or cDNA labeled with fluorescent dyes is hybridized to the chip. Then using a laser system, the presence of fluorescence is checked; the sequences and their quantities in the sample are determined.
7) Use of Next Generation Sequencing (NGS) allows comprehensive description of germ-line DNA, analysis of somatic mutations and RNA profiles in naturally occurring tumors, systematic analysis of microbiomes, etc. The resulting data can help in the identification of novel hereditary syndromes, molecular targets for cancer therapy, tumor-specific diagnostic markers, etc. But this technique would be expensive to work with.
Best Wishes.
  • asked a question related to Real-Time PCR
Question
3 answers
TO detection from two single nucleotide polymorphism (SNP) in one gene, and determine the accurate hypothyroidism in adult
Relevant answer
Answer
What methods you use depend on factors like how many samples and the distance between the snps, how many negative results you expect and whether you will end up sequencing for confirmation after finding the snp. So if your snps are within a single sequencing amplimer and samples are few and the changes are common then sequencing from the start may be good. If you have many samples to test but few at the end to sequence then ARMS screening followed by sequencing interesting detected samples works
  • asked a question related to Real-Time PCR
Question
26 answers
Hi All
If you take a look at the image, the upper virus X assay (for Rotavirus) is giving off a lot of fluorescence at the end of the run (circled) which is clearly not real (we know these are Rota negative samples), this is affecting the baseline setting making the positive samples look like they are giving off poor signal.
The lower figure for the virus Y assay (for Norovirus) shows what the assay should look like but is just an example targeting a different virus
Is this primer/probe degradation and if so why don`t I ever see this is any of my other assays
Many Thanks
MArk
Relevant answer
Answer
I work in Covid laboratory and for 2 years we barely experienced this kind of curves (to be precise 4-5 times in total). However, recently, we faced this late raise curve that 99% appears with one gene (E gene or N gene). During the last two months (Jan and Feb 2022) we faced an outbreak with new COVID variant Omicron. therefore with had to verify which curves are real presumptive and this added to our work load greatly. Some people here asked if someone tried using different extraction method or changing lot number so I am writing this to answer those with these Qs, in case they have not solve it yet. In one case it was primer probe problem. We changed the lot number and the problem solved (see the attached foto). In other cases we found out it happens randomly and doesn't follow any logic. We changed extraction method and we also compared a set of samples with two different extraction methods and saw the curve happened randomly in both conditions.
Briefly, I think it is more probe degradation issue than anything else. Keep in mind Ct under 40 is consider presumptive and can be the onset of infection. So in our lab, we put under investigation all the late raises with one gene and will be re-swab 24-48 hrs later. Statically, we found out around 10 percent of cases turned presumptive with further investigation.
I hope this help. Also if anybody has input on this subject I would be happy to hear your suggestion.
PS.: attached foto is an example of a probe lot number that was not good (degraded probe) in black circle and a random late raise after 35 cycle that only appreared in one gene (N1 gene) circled in blue. The red is positive control gene.
  • asked a question related to Real-Time PCR
Question
3 answers
I'm studying on 5 miRNAs under chilling stress in tomato plants. I would like to know if the concentration of the cDNA used for first stand sample should be the same as the concentration of cDNA used for the main samples on the plate? For example if we have diluted our cDNA 3 fold for the main samples, should we dilute the cDNA that is used for the first standard sample? Should I dilute a diluted cDNA 4 times to prepare 1:10, 1:100, 1:1000, 1:10000 standard samples? Or I should use undiluted cDNA initially ? Thanks for your time and consideration
Relevant answer
Answer
No, you CAN'T extrapolate for qPCR data. Your data points have to be within the linear range of your curve (interpolated).
As Leonid Ilchuk said, you'll have to figure out empirically if any dilution is needed for your particular genes of interest in your samples.
  • asked a question related to Real-Time PCR
Question
3 answers
In the kit guide, it is mentioned that it is recommended to dilute the cDNA 3 fold before making a real-time PCR reaction setup.
Can I add 20 ul of nuclease-free water to 10 ul of cDNA to dilute 3 fold? Or I should add 90 ul of water and then I should take 10 ul of the final solution and add 90 ul of water to it again to make a 3-fold diluted cDNA?
Moreover, I am going to dilute cDNA for standard mix synthesis to use them in the real-time plate vessels. (S1,S2,S3,S4,55). I would like to know for diluting 1:10, 1:100, 1:1000, and 1:10000 of cDNA should I do serial dilutions?
Relevant answer
Answer
take your 10 μL of the stock solution and dilute with diluent to a final volume of exactly 30 μL( add 20 ul of diluent). This will yield 30 μL of diluted solution and results in a final dilution factor of 3. This represents a 3-fold dilution. It may also be said that the stock solution has been diluted 3 fold.
  • asked a question related to Real-Time PCR
Question
4 answers
Hi,
Is there any database to identifies which protein isoform is expressed in a particular tissue?
I want to design primers for real-time PCR on cartilage samples. I need to be sure which mRNA variants are specific for cartilage.
Relevant answer
Answer
Hmm, hard to say...
It seems that isoform uc002bna.2 (http://genome.ucsc.edu/cgi-bin/hgGene?db=hg19&hgg_gene=uc002bna.2&org=human) is expressed in most of the tissue types (see the image in the attachment) ... but there are no data for cartilage tissue type specifically
So you may risk it and proceed further with primer design, or you can try to find expression of your isoform in some cartilage-specific public RNA-seq data (laborious)
Martin
  • asked a question related to Real-Time PCR
Question
10 answers
We have a recurring problem in our routine.
After RNA extraction, we performed reverse transcription of the samples and a negative control with water (Blank). We proceeded with a pre-amplification from this cDNA, including the Blank. We consecutively do the real-time PCR of the pre-amplified samples together with the Blank that "run" along, in addition to using as well as water instead of the sample.
What is happening is that Blank is amplifying (in late cycles), in which case there is no sample in the reaction, so it was not meant to amplify. We have already tested reagent by reagent. That well that we make with water in real time does not amplify, so we discard water contamination.
Do you have any idea what it could be? Have you been through this?
Relevant answer
Answer
Are you running a melt curve at the end (if not, you really should).
"Amplifying late cycles" is a hallmark of primer dimers/mispriming: since the only real difference between your water control and your "blank" is that your blank also contains all the RT reagents (including the primers for that reaction) and the primers used for pre-amplification, I would ascribe "primer dimers" as the most likely source, here. The more primers you mix together, the greater the chances two of them will anneal to each other slightly, and it only takes one aberrant binding event to create a viable amplicon for subsequent cycles.
A melt curve helps distinguish real amplicons (melt peak ~78-85, identical for all wells) from primer dimers (melt peak ~70-75, usually fairly broad), and from mispriming (lots of peaks, messy trace).
  • asked a question related to Real-Time PCR
Question
3 answers
Hello,
I am putting together a standard curve where the concentration of the plasmid is 0.167ug/ul, plasmid bp is 5821, and gene of interest is 1790 bp.
For the plasmid I get : 2.618×1010 copies
For GOI: 8.513×1010 Copies
My question is should I use the plasmid or GOI copy # when developing my standard curve?
I've used this site for my calculation
Copy number calculator for realtime PCR | Science Primer
Relevant answer
Answer
Since you are using the plasmid to make the standard curve, you would use the bp of the plasmid.
Double-check your math before you start, be very careful with your measurements, & mix thoroughly for each dilution.
I'd also suggest that you aliquot the standards into single-use batches & freeze them all. That way you can avoid issues from degrading your standards with multiple-freeze/thaw steps.
Good luck!
  • asked a question related to Real-Time PCR
Question
5 answers
Normally Realtime PCR also known as quantitative PCR (qPCR) is used for investigation gene expression, in which case Reverse Trasncriptase-PCR is used,starting from RNA. Whenever realtime PCR is discuessed, first thing that comes into mind is gene expresseion. My question is what is its role in detection of a mutation? Is it the right thing to use it for mutation detection starting the reaction with genomic DNA? I have done an experiment where I have used qPCR technique for detection of a mutation in whole blood. Genomic DNA was used and primer sets targeting the wild type and mutant sequence were used.
Thanks
Relevant answer
Answer
Yes! Real-time PCR, a hybridisation-based method, has become widely used for mutation detection. The systems are flexible with a number of different probe systems that can be used and there is additional flexibility in the design of the probes. Saima Qayum
  • asked a question related to Real-Time PCR
Question
16 answers
I would like to study correlation between four transcripts (fold changes of mRNA expression) at different time intervals (5 time points). How can I perform this analysis?
Relevant answer
Answer
Try Correlation matrix on R Programming. Try corrplot( ) package in R
  • asked a question related to Real-Time PCR
Question
3 answers
Suppose, I have a treatment group and a control group. I have measured the relative expression (delta delta ct) of two genes by real-time PCR analysis. Both the two genes are analysed from the same samples and the same reference gene was used. In this case, if one gene has higher relative expression than another gene, can I say that gene is highly expressed than the other gene. My question is if the relative mRNA expression from two genes can be compared? another question is, is it ok to produce a heatmap from relative expression data of different genes?
Relevant answer
Answer
Thanks a lot for your answer and suggestion professor Iman Hassan Ibrahim and Dr Abhijeet Singh
  • asked a question related to Real-Time PCR
Question
5 answers
Hi there
I'm using degenerate primers for both Real-time and PCR. Although these degenerate primers work well in Real-time PCR, I can not detect the 100 bp size bond that I'm looking forward it in my Gel electrophoresis after PCR. Does anyone have been faced with such a problem? Any suggestion?
Thank you for your attention and time.
Relevant answer
Answer
If you have amplification in qPCR, with using the qPCR product in gel electrophoresis, you must get corresponding band. If you are not getting it, there is something wrong in your electrophoresis protocol. I don't think primers are a problem in this case.
  • asked a question related to Real-Time PCR
Question
4 answers
Hi everyone!
I want to use an internal control for a real-time PCR method for the detection of DNA viruses. This control would allow me to detect PCR inhibitors. The protocol is based on TaqMan chemistry. I have the primers and the probes, but I don't know how to develop an internal control. It would be really great if someone could help me with this.
Thank you!
Relevant answer
Answer
Hi Erina Noé Seiler, as Salma Aktar said, you can use β-actin or GAPDH as IPC to control inhibitions. You will find PCR sets for both in the following paper: https://gene-quantification.de/overbergh-1999.pdf
  • asked a question related to Real-Time PCR
Question
2 answers
We extract DNA with many kits but the concentration is different in each measure, until zero.
Relevant answer
Answer
We use Qiagen extraction nuclei acid kit.@
  • asked a question related to Real-Time PCR
Question
7 answers
Dear Researchers,
I did Real-time PCR for some miRNAs for (two groups ((responders and non-responders)) to chemotherapy). I had the values of CT values, delta CT, delta delta CT values and finally fold change. I want to do a ROC Curve for them to determine the sensitivity, specificity, PPV, NPV, AUC, and a cut-off value for the test for each marker.
I am using Graphpad prism 7: (Analyze> column analyses> ROC Curve).
I want to know if we put CT values for each marker or fold change and why? Is there any data I have to put beside real-time data?
Moreover, I want to do ROC Curve for combined miRs after the individual ones. How can i do this?
Thank You:)
Relevant answer
Answer
Hi Yusran, first of all, you need to make sure you have a correct and stable normalizer for the miRNA data.
for combination ROC curve, you need to use logistic regression model and picking a correct youden index later according to the sensitivity and specificity value, as a cutoff for your result.
hope it helps!
  • asked a question related to Real-Time PCR
Question
3 answers
Hello everyone
I am currently studying the effect of a particular drug on the expression of a number of genes in the hypothalamic tissue in a number of samples.
After I designed the primers, taking into account the necessary standards, I achieved BLAST for them.
When I made several set up Real-time PCR runs I had a problem of getting either one peak (product), two Peaks (product and Dimer-primer), or one Peak (Dimer-primer) with different samples
Although I raised and decreased both Tm, primers, and cDNA concentration without solving this problem.
In addition, the b-actin primer gave me one peak (product) without any problem with all samples
Could the expression level of the studied genes be related to the extent of the appearance of these three cases?
Relevant answer
Answer
Dear Muhammad Haddad how are you doing?
Primer dimer is most likely related to the primers instead your lower expression. Check the complementary of each primer pair. In addition, the melt temperature is also important in this case. Each primer pair should be design in the same TM temperature. If you are using two-step qPCR reaction, the annealing and amplification occurs at the same temperature for standard 60°C. check your reagents data sheets.
Also, you can perform a conventional PCR reaction and run an agarose gel to access size of the band of primer dimer and your product.
In addition, there is a chance that could be genomic DNA contamination in your cDNA sample. In qRT-PCR analysis, your primers should be design in exon-exon junction or in different exons flanking a big intron. This avoids genomic DNA amplification.
Good luck in your research.
  • asked a question related to Real-Time PCR
Question
2 answers
It seems to me like the 2009 method is preferable, but many recent papers use the 2002 protocol. Is there any reason for this?
Relevant answer
Answer
But, the 2009 procedure is very precise than 2002. We should not sacrifice accuracy for simplicity.
  • asked a question related to Real-Time PCR
Question
6 answers
When performing Viral RNA extractions I perform a negative extraction using water. This is then run as a negative control for my RT-PCR reaction. Recently this control has been showing up RNAse P positive. In addition, when I run the same extract negative sample in duplicate on my PCR plate, sometimes I get RNase P in only one of the duplicates. What is happening? It only ever happens in the Extraction Negative, it never occurs in my PCR negative control.
This issue is occurring for multiple technicians.
Relevant answer
Answer
I am experiencing the same problem. I think it is not contamination but something with the oligo design.
  • asked a question related to Real-Time PCR
Question
3 answers
How can we use traditional pcr to detect corona? Is there an acceptable method of the World Health Organization? What is the last CT obtained from the real-time PCR method that can be seen on agarose gel electrophoresis ?
Relevant answer
Answer
Hi,
I have attached an article in which the protocol explains the use of conventional PCR for the detection of SARS-CoV-2.
  • asked a question related to Real-Time PCR
Question
3 answers
I have performed high-resolution melt analysis using KAPA HRM FAST mastermix; PCR conditions were the same for both experiments - Image 1 and Image 2. However, there is a lot of variability in the two consecutive experiments. I haven't changed any parameter between the two experiments. (Image 2 contains additional samples).
Can anyone point out what I could be the possible reasons for variation in results.
Thank you.
Relevant answer
Answer
Hi
In addition to the above suggestions, the observed problem may also be related to the quality of the RNA in different samples, or cDNA synthesis efficiency. So, compare the integrity of the RNA samples, and the profiles of the reference genes between samples.
  • asked a question related to Real-Time PCR
Question
4 answers
I performed a real-time PCR analysis on cDNA samples deriving from one microalgal species grown under different conditions.
In order to adjust the expression-fold changes on the base of the amplification efficiency, I used one condition (control) to prepare serial dilutions and obtained a calibration curve for each target gene.
I do not know yet if the type of sample significantly influences the efficiency (for a temporary shortage of reagents) but from what I read usually researchers focus just on efficiency variation depending on the primer pairs.
Is this enough or I should do the same also for the rest of the conditions? And if yes, how the Pfaffl formula would change in order to consider different efficiencies per target AND per condition?
Relevant answer
Answer
Hi David
with same primers, kits and conditions your samples shouldn't have huge variations, except when samples quality are not homogeneous.
the best way to get proper results is therefore to ensure quality of your samples and your groups quasi equal, and run your RT-qPCR following the rules for the delta/delta Ct engagement.
I include to this message the axle file allowing Ct calculation.
all the best
fred
  • asked a question related to Real-Time PCR
Question
2 answers
Which Real-Time PCR Instrument is better? Those that work with passive reference dye or those that do not work with passive reference dye.
Relevant answer
Answer
There are instruments that use ROX like ABI, others that use fluorescein (Biorad), others that correct by centrifugation (Rotorgene) ... But I think there are diverse criteria to choose an instrument: if is it an open system for the consumable (master mix and plastic support) or not?, the spectrum (multiplexing: nbr of filters), maintenance (lamp ...), Heating block accuracy, Calibration ... and the cost of course. You can get a comparison at this link : https://www.biocompare.com/PCR-Real-Time-PCR/22353-Real-Time-PCR-Thermal-Cyclers-Thermocyclers/
  • asked a question related to Real-Time PCR
Question
6 answers
I am exploring a number of SNPs, unfortunately although the test are ok, I am perplexed because my preliminary results are nearly all against what is reported in the literature.
To confirm my workings, I need to understand if I have done this correctly.
I am using prepared primers from TaqMan for a realtime PCR. This uses two dyes: VIC and FAM, each associated to one allele.
An example of a primer would be the following:
CCAGCGGATGGTGGATTTCGCTGGC[A/G]TGAAGGACAAGGTGTGCATGCCTGA
where the manaul stated that A should be VIC and G would be FAM.
Does this mean that those results with a low Ct on VIC and a high Ct on FAM would be A or G allele?
Relevant answer
Answer
"The SNP alleles are included in brackets, where the order of the alleles corresponds to the association with probe reporter dyes, where [Allele 1 = VIC® dye / Allele 2 = FAM™ dye]."
TaqMan® SNP Genotyping Assays User Guide
  • asked a question related to Real-Time PCR
Question
13 answers
Greetings all,
I've been performing qPCR on an ABI "StepOnePlus" machine for the past year. I produce RNA either by kit or Trizol, perform RT using Quantabio "qScript cDNA synthesis kit", and perform SYBR green qPCR with Quantabio "perfecta sybr green fastmix rox".
I've recently realized that I am not producing standard curves with good slopes anymore. This is true for new primers I ordered based on literature that are supposed to be perfect, as well as for all my most reliable primer sets that I've previously validated (they used to have good R, Good 3.3-/+0.3 slopes, only one melt curve product, no primer dimers or products in NTC wells).
All my curves now shift between 55-and-75% efficincy instead of >90%, in spite of my R^2 values staying great and the melt curves still showing single products.
I've tried:
making dilution curves from fresh RNA from different sources,
using different SYBR mix tubes  and an identical SYBR mix from another lab,
performing the qPCR on other "StepOnePlus" machines, 
changing PCR plates, optical film, a fresh cDNA synthesis kit,
I validated that all the pipettes I use are well calibrated,
I even changed the molecular grade ultra pure water I use for reactions.
Results are always the same: low efficiency, evident by a growing ΔCt from one dilution to the next. 
Can anybody think of a reason that I may have overlooked? something that changed in how I work that I may have not noticed? Something I that thought I controlled for in one of the troubleshooting runs I detailed above but may have done wrong?
We are getting really desperate here...
Relevant answer
Answer
Gil Diamant Hi, Gil! I am having the same sort of issue. What was it in the end for you? Was it the old batch of plates?
  • asked a question related to Real-Time PCR
Question
4 answers
Dear all,
i have a problem and i cant find any explanations:
I isolated murine Spleen cells and put them in the cell culture for 24h, stimulated with rBAFF in Bcell Medium (RPMI with FCS, Hepes, PenStrep and bMercaptoethanol).
I then wanted to do a realtime PCR with our established Taq Man system, as usual. But i had only amplification in few of the produkt for the housekeeping Gene (hprt).
So i tried the same samples with another established system, Quantitect Primer from Qiagen and Sybr Green and i tried several housekeeping genes (Actb, GAPDH and HPRT).
In the end it seems like the Sybr green PCR worked fine as usual, but the Taq Man didnt. I did Taqman PCR quite a lot with our samples, the only difference now is, that these sample were in cell culture before RNA Isolation. What might happen there, so that the Taq man system isnt working but the Sybr green is??
Does someone have an explanation, it is driving me nuts :DD
Thank you!
Relevant answer
Answer
Some inhibitors in the samples could inhibit the amplification and detection by TaqMan RT-PCR but not by the conventional or nested RT-PCR. For example, check this paper which indicates the same problem:
If you have enough concentration of your samples, try to dilute them as much as possible and try again with the TaqMan.
  • asked a question related to Real-Time PCR
Question
3 answers
I am working on developing a real-time PCR assay but I don't know how to set up and/or define the range for CT values which could help me in lower and upper CT values for determining the positive and negative samples of the experiment.
Relevant answer
Answer
The lower Ct is defined only by the need to have some cycles "in the background" so that one can identify the background signal and recognize the beginning of the exponential part of the amlification curve. This will be difficult ifthe Ct is less than about 10 or so.
The upper Ct is usually at around 35, but this may depend on the assay. To get better estimates of the upper bounnd you may generate a standard (could be a PCR product amplified with primers outside the qPCR primers), purify and quantify this product and measure it in qPCR in a dilution series. You would actually need only the dilution containing 0.5 copy (on average).You can expect 39% of reactions give a positive result, and among those 77% will contain only a single template molecule (the rest will contain 2 or more copies). That should give you a good idea what Ct you can expect if there is just a single molecule in the PCR.
A different question is that for the dynamic range of reliable quantification. Here you must define what "reliable" means. Particulrly at the lower concentrations, Poisson variation limits the achievable precision.
  • asked a question related to Real-Time PCR
Question
5 answers
In nucleic acid extraction for realtime PCR, people often use the external control to monitor the extraction efficiency. The external control can be a DNA/RNA fragment or the cell contains plasmid with the DNA control. So, there are 2 kinds of the external control: the DNA/RNA fragment or the cell. I think using the cell is better because it mimics the biological samples, we can extract simultaneously the samples and the control. Why many manufacturers use the DNA/RNA fragment as the extraction control, not the cell itself?
Relevant answer
Answer
Hi Thanh, I believe that the external controls provided by the manufacturers have been reliably validated to be efficiently isolated via the extraction kit/protocol provided. As different kits have varying limitations on several factors that may affect their respective extraction efficacies, for instance the maximum size of DNA/RNA fragments that can be isolated by that particular kit, using alternative external controls may not reflect on the actual efficiency of the nucleic acid isolation procedure that has been perfomed.
  • asked a question related to Real-Time PCR
Question
5 answers
I am planning to perform MTT assay for determining the anti-apoptotic activity of MSC derived exosomes on drug treated neurons. Following are my doubts.
1)to collect exosomes should I allow the MSC's to reach complete confluency in culture flasks, keep them in that state for 48 hrs and then collect the conditioned media.?
2) what dilution range of exosomes (ug/ml) should I use for adding to the 96 well plate or what should be the highest dilution  i should start with ?
Relevant answer
Answer
Hi, recently I started my work to observe the role of Exosomes (derived from glioma cell line) in cell proliferation. Please, could you suggest me, when will I add exosomes with cells? Just after transferring cells for incubation for MTT when cells are floated or after adhesion on the bottom of flask?
  • asked a question related to Real-Time PCR
Question
2 answers
Can anyone help to explain the real-time PCR results? how can I get a curve graph for fluorescence and amplification?
Relevant answer
Answer
With SybrGreen I got good results. But not with probe Cy5.
with the probe, I did not use any dye.
PCR mixture (10): probe(6.25µM): 0.32µl
Forward and reverse primer (10µM): 1µl each
10x buffer: 1µl
dNTP (2.5µM each): 1µl
DNA Taq pol(5U/µl): 0.1µl
water :5.58µl
  • asked a question related to Real-Time PCR
Question
3 answers
I want expert review regarding the performance of two above mentioned real time PCR systems.
Relevant answer
Answer
I used cfx and quant3 quaigen
but mostly cfx .
  • asked a question related to Real-Time PCR
Question
8 answers
Hello everyone,
I noticed few problems when analyzing my qPCR results. The first being that my reference genes have high Ct value, so when I substract them from my samples, the dCt is negative. This results in negative ddCt and high 2^-ddCt values. Which is a wrong representation of the data. What should I do?
Thank you,
Kindly,
Martin
Relevant answer
Answer
Hi! I agree with Katie A S Burnette and Jochen Wilhelm
If you have a good efficiency of your PCR, and your target genes and reference genes have similar efficiencies, there is no problem with negative delta Ct (Cq). Here you have a good reference to understand how to analyze your data: Analyzing real-time PCR data by the comparative CT method doi:10.1038/nprot.2008.73
When I started my early experiments with qPCR, it was very useful.
I also recommend you this paper: The MIQE Guidelines: Minimum Information for Publication of Quantitative Real-Time PCR Experiments DOI: 10.1373/clinchem.2008.112797
If, after all, you’re still not satisfied with your reference genes… You can use another method to normalize the gene expression, like cDNA quantification with Oligreen.
Best Regards,
Clei
  • asked a question related to Real-Time PCR
Question
3 answers
I am trying to optimise a real-time PCR assay for detection of some cucurbit viruses. Can you, please, advisc reference primers to amplify a housekeeping gene in cucurbits of ca 100 bp (Cucurbita, Cucumis...)?
Relevant answer
Answer
I suppose 18S rRNA can be used universally and can be better as an housekeeping gene.
You can check several publications that uses 18S as a housekeeping gene.
  • asked a question related to Real-Time PCR
Question
6 answers
Hello folks.
I have done a pcr using kodfx enzyme (dntp (2.5 mM each, 10xbuffer) and have got the desired product and optimum annealing temp for my primer. Unfortunetly when I try using platinum taq enzyme and its componet(10xbuffermgcl2, mgcl2..), and same primer with same annealing temperature as before, i didnt find any product.
Can you suggest me some troubleshooting ways?
Relevant answer
Answer
A similar thing happened here. It seems like platinum is not as efficient as Phire Hot Start II taq polymerase.
  • asked a question related to Real-Time PCR
Question
11 answers
I have been using 2X SybrGreen mix by Thermofisher Scientific (#F416L) and their recommended volume is 25 ul. I have scaled it down to 15 ul. However, I have read papers where they use 4ul reaction volume. I want to know if I can still lower down the reaction volume in this particular instrument (Stratagene Mx4000P)
Relevant answer
Answer
I scaled it down to 7 ul and it worked fine.
  • asked a question related to Real-Time PCR
Question
4 answers
I extracted RNA from flax seedlings to measure microRNA expression by stem-loop PCR. I have done this many times with the CTAB method and often the 260/280 ratio is good and around 2 but the 260/230 ratio is between 2.3 and 2.4. Is there a problem with this ratio for cDNA synthesis or real-time PCR? Does this ratio have to be between 1.8-2.2? Please guide me if you have experience in this field.
Relevant answer
Answer
Somayeh Hooshyar The 260/230 ratio while quantifying RNA is used to indicate the presence of unwanted organic compounds such as Trizol, phenol, Guanidine or guanidine thiocyanate. Generally acceptable 260/230 ratios are in the range of 2.0 – 2.2. Values higher than this may indicate contamination with those unwanted organic compounds.
  • asked a question related to Real-Time PCR
Question
13 answers
Hi,
I want to use TOPO cloning with my current experiment.
My insert size ~ 3.5 kb  in size. actually, I did overlapping pcr for my insert then I decided to amplify my insert using TOPO as I noticed in the protocol, I must do taq polymerase pcr in order to add A at the terminals and I did that. however, I repeated the TOPO experiment three times but its not working with me!!! any recommendation to take into account please?
* my kit isTOPO® TA Cloning® Kit,intitrogen
thanks
Relevant answer
Answer
Hello Jose
I have a promoter with nearly 3000 bp called CLEC4F. I wanted to clone the whole sequence of this gene using the TOPOTM XL-2 Complete PCR Cloning Kit. I have some questions about this protocol;
1) Please let me know your comments about the designing of primer for this promoter.
2) Then, I should follow the protocol suggested by kit ?
3) I also saw your above comment. The PCR protocol is (based on the kit protocol)
Thanks,
Raheleh Roudi
  • asked a question related to Real-Time PCR
Question
11 answers
Does anyone know about which carrier is better for RNA extraction: LPA (linear acrilamide), tRNA or glycogen? I need to analyze samples with low quantities of RNA, make c-DNA and Real time PCR. I heard that tRNA can inhibit c-DNA synthesis... is it true? maybe someone who has worked wtih these can help me! thanks!   
Relevant answer
Answer
Now I use Polyacryl carrier (MRC) for total and small RNA precipitations, which works well even in recovering minute quantities. In the past, I used glycogen which worked great too.
  • asked a question related to Real-Time PCR
Question
4 answers
Dear friends!
1. Is there any DETAILED algorithms/videos to design primers and probes for real-time PCR?
2. Do probes really give much accurate results and are easier to work with (troubleshooting)?
Will be grateful to any speculations on that point.
Have a great day 😊
Relevant answer
Answer
Hi Anna,
1. The tools mentioned by Bhairavi are commonly used, especially IDT and primerblast. However, if you are lucky enough to be working with mouse, human or rat, IDT sells pre-designed primer probe sets for most targets. https://www.idtdna.com/site/order/qpcr/predesignedassay
These have already been through trouble-shooting and offer very good results from the first run.
2. Yes, Primer-probes are very nice to work with and tend to give a higher degree of precision than SYBR green. (Can do a serial dilution and determine copy-number if you want to.) Troubleshooting is usually minimal. This is especially true with the pre-designed assays.
Good luck!
BWE
  • asked a question related to Real-Time PCR
Question
3 answers
I have carried out Real-Time PCR in Applied Biosystem 7500. I mistakenly used relative quantification option instead of absolute quantification option. So I cant get Ct Value. But I have typical curve. There is any possibility to calculate the Ct value using the same hyperbolic curve in same software. I have limit of reagents only. Plz help me.
Relevant answer
Answer
I must agree with John Hardy Lockhart .
Do check if you enabled the Ct-value to be shown in your wells (software).
See figure below.
Good luck with your experiments and feel free to contact me if you have any other problems with the ABI software.
  • asked a question related to Real-Time PCR
Question
9 answers
I want to do an absolute real-time PCR for quantifying the expression of 2 different genes. is it necessary to consider a reference gene? what is its role?
Relevant answer
Answer
The reference gene serves as a loading control. Normalizing to a reference gene expression improves the accuracy and precision in cases where you may have some variance in sample preparation, RNA isolation, or cDNA synthesis. I recommend to always measure one (or better even a couple of) reference gene(s), unless you are absolutely sure that you don't have and variance in sample preparation/RNA isolation/cDNA sysnthesis, and if you don't even want to cross-check that everything really works as wanted/expected.
  • asked a question related to Real-Time PCR
Question
3 answers
All components are freeze-dried. We use AB7500.
compare the liquid state,  the other channels' noise signal of freeze-dried components are very high when we only detect one channel
Relevant answer
Answer
Hi Jing,
Have you solved the problem? I have a similar problem in one of the multiplex PCR involving FAM, Hex and Rox channels. Among them Hex is creating troubles with either false positive and/OR giving non-sigmoidal curve shapes similar to what you have shown in your question. Please help. Thanks
  • asked a question related to Real-Time PCR
Question
4 answers
Hello everybody,
In my q-PCR results, I got negative delta Ct values for some genes (where the GOI expression is higher than the reference gene).
In this case, should I repeat the q-PCR runs for the GOI with diluted samples to obtain positive delta Ct? or is there any analysis method that is appropriate with such cases?
Thank you
Relevant answer
Answer
Uh...no. This is fine.
Subtracting ref gene Cq values is a log-space operation, equivalent to division in linear terms, exactly as you should do with a reference.
All you're actually interested in is the relative differences between your samples, for your GOI, and those differences are, as noted, relative: it doesn't really matter what the actual expression level is.
If, after normalisation, two samples are different by four cycles, that's a 16-fold difference. It doesn't matter if one sample is "2" and the other "6", or if one is -"2" and the other "2", or one is "-4" and the other "0", or whatever: it's the difference between them which matters.
If I'm honest, I find that dCt/ddCt methods usually cause more confusion than anything else, because many people are uncomfortable with log-space operations and data (and also, dCt methods assume identical efficiency, which is not necessarily the case).
I typically recommend transforming everything into linear values (relative quantities, RQ) and then just dealing with those instead.
For this, you take all the Cq values for a given gene, and find the lowest value (highest expression).
Then for each sample, the equation is
Sample RQ = efficiency(lowest Cq - sample Cq)
Where efficiency is extent of doubling at each cycle (so 100% efficient = 2.0, 95% efficient = 1.95 etc).
What this does is render all of your sample values linear, relative to the highest expression, on a scale of 0-1.
You then do the same for your reference genes, determine the geometric mean of those to generate a normalisation factor, and then divide all your GOI RQ values by the normalisation factor (because we're in linear terms now, so normalisation is a division operation).
And now you have normalised, linear data.
If one sample is 0.1 and another is 1.6, that's a 16-fold difference.
This way you don't have to worry about negative data and what it means.
  • asked a question related to Real-Time PCR
Question
5 answers
I have some qPCR data that I want to analyze with LinRegPCR, but I can't find the way to export the data in a format recognized by LinReg.
Relevant answer
Answer
Hi Irfan and Laura. i can't do it. The software show me a message "Could not open text file for exporting data". How can i do to solve it?
  • asked a question related to Real-Time PCR
Question
18 answers
Is it possible to run a real-time PCR product in agarose gel electrophoresis? Would bands be visualised?
Relevant answer
Answer
Thanks for you response